I could do with some help on wiring in my load cell 

I have and s type 300kg cell 

ther is 5 wires come from it 2 say input +\- 

2 say output +\- 

1 says gnd 

any help on Were to install these is greatly appreciated 

 
Posted : 22/11/2019 9:35 am
(@admin)
Posts: 1457
Member Admin
 

Hi,

The good news here is that it you are not going to damage anything if you get it wrong. 

Gnd should go to the shield (I expect), input +/- goes to 0V and 5V (does not matter which is which) and output +/- goes to Signal+ and Signal-. Again it does not matter which is which. Normal is that compression gives a positive number and stretch gives a negative number, so you can swap Signal+/- if it is not the case. But this is anyway taken care of in the calibration, so it is not really necessary. 

 

 